Follow along for an enriching discussion with our CEO, Peter Herweck, and industry leaders for a captivating Global Keynote at #InnovationSummit Paris 2024. Watch recording and join the #ImpactMaker movement!️
IMD-IT-S100-H Compliance with IEC 61558-2-15
Solutions for your Business Forum
No matter if your business is in the retail sector, food manufacturing, or healthcare, you'll find resources and a community of experts to discuss with on content tailored to your industry sector. Subscribe today.
Send a co-worker an invite to the portal.Just enter their email address and we'll connect them to register. After joining, they will belong to the same company.
You have entered an invalid email address. Please re-enter the email address.
This co-worker has already been invited to the Exchange portal. Please invite another co-worker.
Please enter email address
Send InviteCancel
Invitation Sent
Your invitation was sent.Thanks for sharing Exchange with your co-worker.
Link copied. Please paste this link to share this article on your social media post.
Posted: 2021-07-0904:07 AM. Last Modified: 2022-06-2401:21 AM
IMD-IT-S100-H Compliance with IEC 61558-2-15
Dear My Colleagues,
I am checking the IMD-IT-S100-H (IT Medical Transformers - Asia Range) specifications for medical application.
1. IMD - IT - S100H complied with IEC 61558-2-15: 2011 standard.
2. Clause 18.102 IEC 61558-2-15: 2011 standard mentions: The leakage current through the protective earth conductor from the enclosure to earth, shall not exceed 0,5 mA when measured at room temperature under no-load condition with the transformer supplied at the rated voltage and the rated frequency.
3. But on the transformer spec, leakage current item it shows that " The leakage current is less than 3.5mA between primary and secondary, primary and bracket, when input 230V/ 50HZ"
With the specification in [3], customer understands that "The leakage current < 3.5mA" and it didn't comply with item [2].
It is so confuse for me and also customer.
My question: Can we confirm with customer that: “our isolation transformer for medical locations IMD-IT-S100-H has leakage current through the protective earth conductor from the enclosure to earth is not exceed 0.5mA as per IEC 61558-2-15 standard ?"
If yes, do we have any update document to show it.