Hi all,
I'm not sure if this is the correct forum but...
I am looking for an UPS to run a mass spectrometer (900 VA) and a PC workstation that are experiencing issues due to brief sags and surges caused by an associated instrument turning on. The mass spectrometer is very sensitive to even millisecond fluctuations in power and will shut down as a result. Over time the constant shutting down blows the smart card in the spectrometer.
I'm looking for a UPS that will work for this application. Any advice?
Thanks.
Posted: 2021-06-29 12:59 AM
This reply was originally posted by Jonathan on APC forums on 9/22/2010
Hello Quinton,
You might want to consider our Smart-UPS Online --> [Smart-UPS Online Product Information Page|http://www.apc.com/products/family/index.cfm?id=163]. They offer zero transfer time and can accept a wide voltage range which is good if you're having frequent sags or surge. If you want to know more about the different UPS topologies you can view our white paper here --> [White Paper #1|http://www.apcmedia.com/salestools/SADE-5TNM3Y_R7_EN.pdf].
Posted: 2021-06-29 12:59 AM
This was originally posted on APC forums on 6/29/2011
Dear JonPro
We have problem SURT6000XLICH UPS. PLease could you see log the
2011-06-29 12:12:19 UPS: Passed a self-test.
2011-06-29 12:12:15 UPS: An inverter fault no longer exists.
2011-06-29 12:12:10 UPS: An inverter fault exists.
2011-06-29 12:12:06 UPS: An inverter fault no longer exists.
2011-06-29 12:12:03 UPS: Refused a self-test; UPS is overloaded.
2011-06-29 12:11:59 UPS: An inverter fault exists.
2011-06-29 12:11:55 UPS: An inverter fault no longer exists.
2011-06-29 12:11:50 UPS: An inverter fault exists.
2011-06-29 12:11:47 UPS: Passed a self-test.
Please could you send me about it. Please could you consult me about it.
Regards
Azaraa
Posted: 2021-06-29 12:59 AM
This reply was originally posted by Jonathan on APC forums on 6/29/2011
I have responded on another thread that you have created Azka.
Posted: 2021-06-29 12:59 AM
This reply was originally posted by Jonathan on APC forums on 9/22/2010
Hello Quinton,
You might want to consider our Smart-UPS Online --> [Smart-UPS Online Product Information Page|http://www.apc.com/products/family/index.cfm?id=163]. They offer zero transfer time and can accept a wide voltage range which is good if you're having frequent sags or surge. If you want to know more about the different UPS topologies you can view our white paper here --> [White Paper #1|http://www.apcmedia.com/salestools/SADE-5TNM3Y_R7_EN.pdf].